Thank you all for your insights and information.
I think the best way forward is simply to clarify in the definitions
that the cell_measurement_temp/pressure items would only differ
(slightly) from _diffrn_ambient_temp/pressure where the cell was
determined separately from the diffraction pattern. If these items are
absent their default values can be taken from the corresponding
_diffrn_ambient items. Making the reverse deduction can be a matter of
policy for CIF curators.
If there are no objections, I will make the updates to the core
dictionary. You can follow along and comment here:
https://github.com/COMCIFS/cif_core/issues/294
On Tue, 18 Jan 2022 at 20:19, Spek, A.L. (Ton) <A.L.Spek at uu.nl> wrote:
>
> James,
> A good point.
> IUCr/PLATON/checkCIF sends out an A-level ALERT for this. Referees should have picked this up.
> Older SHELXL versions put in a default value for those values in the CIF they generated (Room Temperature)
> A quick check of the CIF's for all structures published in the 2021 issues of Acta Cryst. E77 shows that there are no inconsistencies there.
>
> Best wishes
>
> Ton
>
>
> Op 18-01-2022 06:47 heeft coreDMG namens James H via coreDMG <coredmg-bounces at iucr.org namens coredmg at iucr.org> geschreven:
>
> Happy New Year everybody.
>
> Here is our first issue for the year:
>
> At first glance it would appear to me that _diffrn_ambient_temperature and
> _cell_measurement_temperature are describing the same thing. There may
> have been a difference in the past when using point detectors and the
> cell was determined separately to the data collection, but for
> anything collected on an area detector I would have expected these
> values to be the same. However, I've noticed from looking at around
> 10000 files from the COD that, of the files that have both data names,
> around 1% have different numbers. I'm wondering what this could
> possibly mean, especially when the two numbers are very different e.g.
> 150K and 296K. Given that the cell parameters are necessary to obtain
> absolute distances from fractional coordinates, if the diffraction
> data are collected at a different temperature to the cell measurement,
> then surely these atomic distances would be incorrect? How do the
> database curators handle this? How are software authors interpreting
> these two data names for output or input?
>
> If you agree that there is no way these numbers should be
> significantly different, suggestions on how to fix the situation are
> welcome. Personally I would deprecate _cell_measurement_temperature or
> state that it is only used when the cell is measured separately to the
> main diffraction data set.
>
> James.
>
> --
> T +61 (02) 9717 9907